Question1 Serum glucose determinations are done on samples from a control pool e
ID: 51371 • Letter: Q
Question
Question1
Serum glucose determinations are done on samples from a control pool each day for 30 days. Quality control statistics are then calculated. The mean is 100 mg/dL and the standard deviation is 4 mg/dL. These statistics are used to label a quality control chart for the mean and +2 SD. Samples from the same control pool are then assayed each day along with patient samples, and control results are plotted on the chart. Results for the first 10 days are as follows: see figure below
On day 7, the control results are more than two standard deviations from the mean. The first thing you would do is:
Make new reagents and run the control and patients for that day again
Rerun the control with the same reagents
Discipline the technologist who was responsible for the unacceptable work
Call the company to repair the instrument
Question2
Serum glucose determinations are done on samples from a control pool each day for 30 days. Quality control statistics are then calculated. The mean is 100 mg/dL and the standard deviation is 4 mg/dL. These statistics are used to label a quality control chart for the mean and +2 SD. Samples from the same control pool are then assayed each day along with patient samples, and control results are plotted on the chart. Results for the first 10 days are as follows: see figure below
Using 95% confidence limits means that 95% of the control values would fall between:
96 and 104
92 and 108
88 and 112
Can't tell from the information
Question3
A patient's blood urea nitrogen (BUN) was 50 mg/dl on Day 1, and the control serum for BUN was within 2 SD of the mean on that day. On Day 2, the same patient's BUN was 70 mg/dL; again, the control value was acceptable. Should the physician interpret this increase in the BUN as real pathologic change in the patient? (The only control level available for this test has a mean value of 20 mg/dL and a standard deviation of 2 mg/dL.)
True
False
Question4
Before the first morning run of patients are placed on a chemistry profiler, the operator notices that the control values for creatinine are 1.1 and 5.8 mg/dL respectively for level I and II. The +/- 2SD range for level I is 0.9 - 1.1 mg/dL and for level II is 5.3 - 5.5 mg/dL. The operator should:
Report patient creatinines
Do not report patient values until the QC problem has been resolved
Report patient values after recalibrating creatinine
Report patients who have plasma creatinine within the normal range
Queastion5
When measuring enzyme activity, if the instrument is operating 5 degrees C lower than the temperature prescribed for the method, how will the results be affected?
Lower than expected
Higher than expected
Varied, showing no particular pattern
All will be clinically abnormal
Question6
Which of the following methods of measuring enzyme activity utilizes an increase in absorbance at 340 nm as an indicator reaction?
"Forward" LD reaction
"Reverse" LD reaction
AST(GOT) reaction
ALT(GPT) reaction
a.Make new reagents and run the control and patients for that day again
b.Rerun the control with the same reagents
c.Discipline the technologist who was responsible for the unacceptable work
d.Call the company to repair the instrument
Explanation / Answer
Question 1 - Answer a. Since the values are higher than the control we could guess that there will be some problem with the reaction. If the same results persist we can go for other options.
Question 2 - Answer a. Since the standard deviation is 4 and 95% confidence level by 0.95 the values will fall only in the range of 96 to 104.
Question 3 - Answer True. Since the standard deviation is only 2 mg/l the values are acceptable and the increase in the BUN is due to pathological condition.
Question 4 - Answer b. He should not report untill the qc problem is solved.
Question 5 - Answer a. The results will be lesser than expected since the enzyme does not work at its optimal conditions.
Question 6 - Answer c. AST(GOT) reaction since the NADH increase results in absorbtion at 340 nm