You are developing a procedure for determining traces of copper in biological ma
ID: 475508 • Letter: Y
Question
You are developing a procedure for determining traces of copper in biological materials using a wet digestion followed by measurements by atomic absorption spectrophotometry. In order to test the validity of the method, you obtain a NIST orchard leaves standard reference material and analyze this material with your procedure five times and obtain a mean of 12.1_7 ppm with a standard deviation of o 08_5 ppm. The NIST sample is listed as 11.6 ppm. Does your method give a statistically valid value at the 95% confidence level?Explanation / Answer
For the given data
analysis of copper
number of reading = 5
mean = 12.17 ppm
standard deviation = 0.085 ppm
at 95% confidence level the value range would be = 12.06446 and 12.27554
So, the NISTsample listed as 11.6 ppm does not fall in the given range and method does not give a valid statistical value.