You are developing a measurent system to measure the output force from a 50 hp s
ID: 2073459 • Letter: Y
Question
You are developing a measurent system to measure the output force from a 50 hp small engine. The main instrument to be used is a load cell. Go to the following Omega webpage: 1. https://www.omega.com/ppts/LC101 html If you are expecting the force you need to measure to be between 20 lb and 45 lb: a) Please select the most appropriate load cell to be used and provide the model number. Note: the "cable" model is required. b) Using the specifications for the load cell, determine the overall instrument error (in pounds) that is expected from your chosen load cell.Explanation / Answer
a) we are epecting a force in between 20 lb to 45lb
so in this range of force we have only one load cell available and
model is LC101-25, it is a cable connection model.
b) the specifications of the model
Exitation - 10 VDC, 15 VDC max
output- 3 mV/V +- 0.0075mV/V
Linearity- +- 0.03% FSO
repeatability +- 0.01% FSO
operating temperature range - -40 to 90 degree centigerade
compansated temperature range - 17 to 71 degree centigrade
safe overload- 150 % of capacity
ultimate overload -300 % of capacity