Determine the minimum number of theoretical plates, N, required to achieve basel
ID: 580590 • Letter: D
Question
Determine the minimum number of theoretical plates, N, required to achieve baseline separation with a resolution that is desirable for quantitative analysis of two compounds with an adjusted relative retention, a, of 1.048 and retention factors of 5.58 and 5.85. Incorrect. A minimum desired resolution for quantitative analysis is considered to be 1.50. Use the following relationship to determine the number R-- Number - N- 934 plates of theoretical plates, where R is the resolution, is the adjusted relative retention, and k2 is the retention factor of the later eluting In an attempt to increase the resolution of the separation, a researcher decreased the column length. True or false, decreasing the column length will increase the resolution of the separation? False O True @previous Give Up & View Solution # Try Again O NextExit- | | LExplanation / Answer
For Number of theoretical plates (N)
Using,
R = (sq.rt.(N)/4)[(alpha-1)/alpha][k2/(1+k2)]
with,
R = 1.5
alpha = 1.048
k2 = 5.85
we get,
1.5 = (sq.rt.(N)/4)[(1.048 - 1)/1.048][5.85/(1+5.85)]
N = 23328 plates
Decreasing the column length will increse the resolution of the separation,
False
Longer run time increases the separation between the peaks and hence resolution increases with column length.