ADCs digitize at different rates. What conversion rate is required if a chromato
ID: 3280746 • Letter: A
Question
ADCs digitize at different rates. What conversion rate is required if a chromatographic peak is to be sampled and digitized 25 times between the first positive deflection from the baseline until the peak returns to the baseline? The total baseline-to-baseline time is (a) 20 s and (b)1 s. According to the Nyquist sampling criterion (see Section 5C-2), a signal must be digitized at a rate at least twice that of the highest frequency in the signal to avoid a sampling error. If a particular 12-bit ADC has a conversion time of 5 s, what is the highest frequency that can be accurately digitized while satisfying the Nyquist criterion?Explanation / Answer
given, the signal has to be digitized 25 times from the first p[ositive deflection untill the peak returns to baseline
a. total baseline to baseline time = t = 20 s
now, for digitisation 25 times in this interval
dt = t/25
so conversion rate = 1/dt = 25/t = 25/20 = 1.25 Hz
b. for t = 1 s
dt = t/25
rate = 25/1 = 25 Hz
a 12 bit ADC has conversion time of 5 micro s = t
so, conversion freq = 1/t
hence this freq must be atlewast three times the signal freq f
so, 1/t = f
highest frequency that can be recorded by ADC accurately = 1/3t = 1/3*5*10^-6 = 66,666.66 Hz