You are charged with analyzing 25 water samples for lead. EPA specifies action i
ID: 480964 • Letter: Y
Question
You are charged with analyzing 25 water samples for lead. EPA specifies action if more than 10% of the samples test at greater than 0.015 mg/L. Additionally, EPA mandates a QA protocol of one Sample Duplicate for every ten samples and one Laboratory Fortified Sample Matrix for every twenty samples. The instrumental protocol calls for a calibration curve of at least four non-zero standards plus a Calibration blank. The calibration curve must be verified before analyzing any samples, after every 10 analyses, and again at the end ofthe anlysis. Prepare a table that lists the sequence in which the samples, spikes, duplicates and standards should be run in order to be compliant with the EPA regulations.
Explanation / Answer
Run samples to make a Calibration Curve, this includes
(ONE blank sample ( no Pb ) and FOUR non zero samples depending upon the range. (here 0.015 mg/L is given so Calibration samples should start from below that value and should be upto slightly more than 0.015 mg/L )
Next step should be to analyze 10 of the total 25 samples.
The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )
Now run any one of the samples again (Duplicate reading )
( from any one of the previously tested 10 samples )
Now test the other 9 of the remaining samples.
The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )
Now take the reading of the one of the remaining sample ( 5 left now )
Now run any one of the samples again (Duplicate reading )
( from any one of the previously tested 10 samples )
Test for the remaining 5 samples.
The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )
STEP 1Run samples to make a Calibration Curve, this includes
(ONE blank sample ( no Pb ) and FOUR non zero samples depending upon the range. (here 0.015 mg/L is given so Calibration samples should start from below that value and should be upto slightly more than 0.015 mg/L )
STEP 2Next step should be to analyze 10 of the total 25 samples.
STEP 3The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )
STEP 4Now run any one of the samples again (Duplicate reading )
( from any one of the previously tested 10 samples )
STEP 5Now test the other 9 of the remaining samples.
STEP 6The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )
STEP 7 Now take the reading of the Laboratory Fortified Sample Matrix. STEP 8Now take the reading of the one of the remaining sample ( 5 left now )
STEP 9Now run any one of the samples again (Duplicate reading )
( from any one of the previously tested 10 samples )
STEP 10 & 11Test for the remaining 5 samples.
The next step is to Calibrate again. So we have to run the Calibrating samples.
( ONE blank and FOUR non blanks [ same we had used earlier to calibrate our system ] )